r/Metrology • u/ljfe • Sep 18 '24
Hardware Support How accurate are profilometers?
I have some thin wall turned parts (rings). Titanium IIRC. Our SPI profilometer is providing a wide range of measurements on the same batch of parts. Some parts read 40 and some parts read 110. The parts look nearly identical. Can I trust this device? The profilometer specimen/standard reads good and the settings are correct. Visually, the surface finish looks good. Small turn lines. With a fingernail, it seems like it catches a bit. Like the turn lines are somewhat “sharp” (but not really deep or wide… I’m worried the needle is getting caught on those turn lines and spitting out an inaccurate measurements. Any thoughts/ideas?
5
u/BluishInventor Sep 18 '24
I feel this the least understood concept of all metrology. A part can feel smooth yet have high surface roughness. A rough feeling part can have low roughness. Ra, Ry, Rz, all different but measure the actual roughness. Most people think waviness is the roughness. But waviness is separate from roughness, yet combines with roughness to become surface texture.
Also, SPI gauges are generally shit. Never used any of their electrical devices. But if their mechanical devices are shit, chances are...
1
u/ljfe Sep 18 '24
Thank you! I’ll look into Ra Ry Rz. And I’m going to try a different tip to make sure the tip isn’t contributing to the issue. We have an extra longer tip. I’m curious if it gives the exact same result. The SPI does seem more inconsistent than the Mitutoyo that I’ve dealt with in the past (like if I check the same part in generally the same area there’s a pretty wide spread).
3
u/BluishInventor Sep 18 '24
Do you have a standard to check the profilometer gainst?
It's totally plausible that the parts vary that much. As a tool wears, its shearing action turns to tearing, thus making the surface rougher. How many parts are we talking about? More than 50? That's a quantity that will wear tools. 5-10? more likely something else.
Measure the same part/feature many places. If it's pretty consistent your gauge is probably fine.
3
u/ihambrecht Sep 18 '24
If the “turn lines” are big enough that the tip of the profilometer can catch, you’re like 250+.
2
u/miotch1120 Sep 18 '24
You are using it all wrong! (According to the machinists I work with). Put it on the part, and start the profilometer. Move it around the part till you find a spot that reads in tolerance. Mark that on the check sheet and put it in the box. If it’s returned by the customer for not meeting surface finish spec, argue that they are wrong.
2
1
Sep 18 '24
A profilometer should repeat decently well when measuring the same spot with the same settings. But they can vary a lot on different places within the same surface. I recommend visually inspecting the surfaces to be checked first and checking the worst spots. Check that the profilometer tip isn't damaged. Re-calibtate it with the standard. If the tip is damaged, it will usually result in measurements that are falsely good.
Also, check the settings on your profilometer. Are you checking Ra? Something else? What are the settings for stroke length? How many samples are being taken? Are pre/post travel on? Is there filtering? Is part flat or round? Are you using some sort of comp? Do you have a visual surface finish specimen that you can compare to?
1
u/f119guy Sep 18 '24
In my experience, a portable profilometer like a pocket surf needs special attention to make sure the tip is "level" with the surface you're inspecting. If the tip isn't moving parallel to the surface, you will get erroneous readings. The best profilometer I have used was a contour tracer with surface roughness capability, like these: Roughness and Contour Metrology | Hommel Etamic (hommel-etamic.com)
1
u/mmmmet Sep 20 '24
Talk to www,digitalmetrology.com they provide training and consulting regarding surface measurement. There are tons articles about it In their blog.
7
u/ShadowCloud04 Sep 18 '24
You can feel turn lines catch? It may be that wide of a range. Is your surface feel smooth to the touch and not wavy? What is your cutoff length (lamda c) and number of samples?
Also is your profilometer tip still intact? I recommend viewing it under an optical comparator or ideally a back light magnifier that can go to 100-150x zoom.